He/N2 carrier gas switching
Posted: Wed Jun 22, 2022 9:20 pm
Are gas filters an issue when switching between He & N2 carrier gas? The following Agilent installation instructions indicate a long purge may be necessary to remove the nitrogen.
"Flush the filter. Set the instrument split vent flow (purge flow) to 400 mL/min, then:
• If using a Moisture Filter in combination with an MS detector, purge the filter for at least 60 minutes. This will flush the filter of dry nitrogen so that it does not appear as a background ion in your MS analysis."
"Flush the filter. Set the instrument split vent flow (purge flow) to 400 mL/min, then:
• If using a Moisture Filter in combination with an MS detector, purge the filter for at least 60 minutes. This will flush the filter of dry nitrogen so that it does not appear as a background ion in your MS analysis."