We are seeing loss of metribuzin of >40% during SPE (EPA 525). All other compounds, including BaP and Perylene-d12 are OK. Diluted standard recovers fine for all compounds. Could it be the extraction solvents, in this case ethyl acetate or dichloromethane? [The DCM we buy is preserved with amylene to prevent formation of radicals that could destroy PAHs]. Has anyone experienced this?
Thank you for your input!
I run these often and have been averaging about 80-85% recovery when spiked at a mid level and 70-80% recovery spiked at the MRL using the same solvent, including preserved DCM.
Do you use the Waters HLB cartridges listed in the method?
How fast do you draw the eluting solvent through the cartridges?
We pull it through until it just starts to drip, then shut off flow for a minute or two, then pull it until all has passed through the sample delivery line to the cartridge then allow to gravity feed until the solvent gets to the top of the packing and stops, then pull through with vacuum slowly until all has passed through. We do this with both ethylacetate and DCM. Pulling through faster seems to give lower recoveries.
When doing solvent evaporation, we also bring it down to about 2ml then wash the tubes with another 2ml of ethyl acetate before continuing down to less than 1ml which also helps with recovery.