Newbee to HPLC

Basic questions from students; resources for projects and reports.

3 posts Page 1 of 1
Hi all..
We have revived a old analytical HPLC (monocratic) system and put a C18-RP column.The eluent flow rate (80:20 MeCn and H20) of eluent is fine without any back pressure problems.
So.. I would like to know the first expirement to be done to check how usefulness of the instrument :idea: .
Pls help ..

what is it going to be used for?

the first thing to do would be to obtain a copy of the column test procedure from the manufacturer of the C18 column you are using and run that procedure to verify that the instrument and column are functioning properly. After that, it depends on your answer to grzesiek's question: what do you intend to use the instrument for?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 89 users online :: 0 registered, 0 hidden and 89 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 89 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry