accuracy validation failure
Posted: Sat Nov 07, 2009 2:50 am
I need advice to troubleshoot an accuracy validation failure of a reversed-phase HPLC assay/degs method for a tablet dosage form. The accuracy experiment design was 6 preparations of drug substance (DS) and known degradants spiked into the excipient matrix (investigation phase validation). The DS was spiked at the 100% level, and the degradants were spiked at the 0.1% level. Accuracy passed for the degs but failed for the DS with two high recoveries of 103.5% and 103.1% (acceptance criteria of 97.0% to 103.0% for DS). The other recoveries were 99.2%, 100.6%, 101.4%, and 102.3%.
I executed the validation several times on a non-GMP instrument and each time with very consistent, passing recoveries between 99.0% to 101.0% before I handed it off to the validation chemist. I also did full characterization of the method consistent with validation requirements of ICH Q2. This included full robustness testing. So far in the investigation, we've ruled out the instrument (including the injector) and any obvious analyst errors (such as pipetting, QS to volume, and weighing).
The method uses quantitation by external standard, and one thing that concerns me is that the standard preparation is not the same as the sample preparation. Standards are prepared by dissolving 30 mg of DS with 50 mL of MeOH / H2O (80:20, v/v), then QS to 100 mL. Samples are prepared by extracting the DS from crushed tablets (100:524 DS/Excipients, w/w) using 50 mL of MeOH. Sonication for 20 minutes, followed by shaking for 40 minutes. QS to 100 mL with MeOH, then pipet 15 mL into 50 mL and QS with MeOH / H2O (80:20, v/v).
The final concentrations of standards and samples are both 0.30 mg/mL, but the solvent composition of each solution is slightly different. The composition of standards is MeOH / H2O (80:20, v/v) and the composition of samples is MeOH / H2O (86:24, v/v). I'm wondering if the method may not be as user friendly as I thought because of these preparation differences.
Any advice is much appreciated.
I executed the validation several times on a non-GMP instrument and each time with very consistent, passing recoveries between 99.0% to 101.0% before I handed it off to the validation chemist. I also did full characterization of the method consistent with validation requirements of ICH Q2. This included full robustness testing. So far in the investigation, we've ruled out the instrument (including the injector) and any obvious analyst errors (such as pipetting, QS to volume, and weighing).
The method uses quantitation by external standard, and one thing that concerns me is that the standard preparation is not the same as the sample preparation. Standards are prepared by dissolving 30 mg of DS with 50 mL of MeOH / H2O (80:20, v/v), then QS to 100 mL. Samples are prepared by extracting the DS from crushed tablets (100:524 DS/Excipients, w/w) using 50 mL of MeOH. Sonication for 20 minutes, followed by shaking for 40 minutes. QS to 100 mL with MeOH, then pipet 15 mL into 50 mL and QS with MeOH / H2O (80:20, v/v).
The final concentrations of standards and samples are both 0.30 mg/mL, but the solvent composition of each solution is slightly different. The composition of standards is MeOH / H2O (80:20, v/v) and the composition of samples is MeOH / H2O (86:24, v/v). I'm wondering if the method may not be as user friendly as I thought because of these preparation differences.
Any advice is much appreciated.