Sudden loss of sensitivity of API5000 in MRM mode
Posted: Wed Nov 04, 2009 8:48 pm
We have an API 5000 running trouble free in the last one and a half year since installation. Last week after routine PM, all of sudden, the sensitivity lost 90% of that before PM.
Interestingly, the sensitivity loss only ocurres to scans associated with collision cell, such as MRMs, daughter ion scan. Q1 and Q3 scan are normal, so it excludes the source or detector problem.
When looking at the PM data, I noticed the ion transmission efficiency (MS/MS), is only 30% of last PM value, just near the lowest acceptable range. I am wondering what could have happened. During the PM, only down to Q-Jet was cleaned. Anyone has some explanation or expereince? At lease let's have some fun. Enjoy and thanks!
Interestingly, the sensitivity loss only ocurres to scans associated with collision cell, such as MRMs, daughter ion scan. Q1 and Q3 scan are normal, so it excludes the source or detector problem.
When looking at the PM data, I noticed the ion transmission efficiency (MS/MS), is only 30% of last PM value, just near the lowest acceptable range. I am wondering what could have happened. During the PM, only down to Q-Jet was cleaned. Anyone has some explanation or expereince? At lease let's have some fun. Enjoy and thanks!