Posted: Wed Nov 04, 2009 7:42 pm
by offroad
Dear friends,
what detector is more suitable for measuring water content?
Thanks
Posted: Wed Nov 04, 2009 9:54 pm
by JGK
Like many forms of analysis it will depend on what type of sample you want to analyze
Solid, Liquid, Gas ?
In short we need a bit more information.
Posted: Wed Nov 04, 2009 11:12 pm
by offroad
Thank you for answering.
My sample is: water and phenols.
They are all dissolved/diluted in water. I would like to know how to quantify water, once I have the following detectors:
Attached to an HPLC:
Refractive index detector - RID
Photovoltaic Diode Array - PDA
Attached to a GC:
Mass Spectometer - MS
Flame Ionization Detector - FID
Thermal Conductivity Detector - TCD
Thank you in advance.
Marcio
Posted: Thu Nov 05, 2009 4:36 pm
by JGK
Are the phenols at trace levels only? i.e. water is the majority of the matrix.
or is it a phenolic solution with trace amounts of water?
I don't think I've ever heard of quantifying "Water" by HPLC or by GC (but I've only been in Chromatography 20+ years).
Moisture Analysis, check out the techniques listed here
http://en.wikipedia.org/wiki/Moisture_analysis
Posted: Thu Nov 05, 2009 6:06 pm
by offroad
Are the phenols at trace levels only? i.e. water is the majority of the matrix.
Yes, water is the majority of the matrix.
I am really new to this subject, so why I can't understand why water may not be properly quantified by a GC (for example).
Thank you for your thoughts.
Marcio
Posted: Thu Nov 05, 2009 6:17 pm
by chromatographer1
I have done water at percent levels and ppm levels.
I suggest the R porous polymer or one of the polar Chromosorb 100 Century series polymers.
TCD is preferred.
Use columns that are not bare metal, but glass or fused silica coated.
Keep oven temperatures above 90°C and inject a minimal amount consistently with a syringe (Chaney adapter) or use a liquid valve injector.
best wishes,
Rodney George
consultant
Posted: Thu Nov 05, 2009 6:27 pm
by offroad
Wow...Thank you.
Posted: Thu Nov 05, 2009 9:23 pm
by JGK
Are the phenols at trace levels only? i.e. water is the majority of the matrix.
Yes, water is the majority of the matrix.
I am really new to this subject, so why I can't understand why water may not be properly quantified by a GC (for example).
Thank you for your thoughts.
Marcio
To be honest this is the first time I've ever heard of anyone wanting to quantify water. Especially when from your answers it would appear you have a water sample with small/trace levels of phenols in it (i.e. it is virtually 100% water).
With this type of sample, I would expect quantitation of the impurities rather than the concentration of water present.
Posted: Thu Nov 05, 2009 9:40 pm
by offroad
To be honest this is the first time I've ever heard of anyone wanting to quantify water. Especially when from your answers it would appear you have a water sample with small/trace levels of phenols in it (i.e. it is virtually 100% water).
With this type of sample, I would expect quantitation of the impurities rather than the concentration of water present.
Well, actually, it's not traces, I believe it's more or less 2-5 % phenols. And there's a few of them that goes from 8 to 15 % phenols, catechol, hydroquinone and water, water is expected to be 70% of the mix.
Posted: Fri Nov 06, 2009 12:49 am
by chromatographer1
Sometimes it is easier to measure water content than to measure each and every impurity to measure water by difference.
Karl Fisher would be preferred but sometimes you are limited by your equipment and personnel.
I have done 2500 ppm water in acetone ACCURATELY +/- 50 ppm and 0-100 ppm water in alkanes +/- 2ppm.
The most difficult part is injecting reproducibly, but I started in 1971 when autosamplers were rare and chemists were plentiful and cheap, as was their time.
Rodney George
consultant
Posted: Fri Nov 13, 2009 11:02 pm
by offroad
Thank you all, very much.
Posted: Sat Nov 28, 2009 9:42 am
by smith2287
hi.
I want to know that chromatography may also depends on the mass of the substances ?
Posted: Mon Nov 30, 2009 12:13 pm
by gstaepels
No, not exactly. In GC the response of the dector to an analyte is the key factor. Therefore, in order to quantify one should use a standard with known weights of the sustances to measure the response of an analyte to the detector.
Many people will use area % and thus neglect the response of the detector. this fine as long as you stay with the same type of samples and use the results relative to eachother. But be aware that these results are not weight %.
Posted: Mon Nov 30, 2009 4:12 pm
by HW Mueller
Gilbert, what is
"the same type of samples" . . . . , and how do you "use the results relative to each other" . . . ?
Posted: Tue Dec 01, 2009 1:40 pm
by gstaepels
With "the same type of samples" I mean a set of samples with the same matrix, in which the components may differ in concentration: e.g. samples from a chemical process, taken at various times.