Hi there, I am having serious problems with active carryover when I run my method on a waters alliance system. I have tried altering solvents, needle wash, everything, but as soon as I inject the sample onto the system the carryover is back, and I cannot diminish it.


I dont have the same problem with Agilent

I have read that Waters systems can have this problem, but I dont know why
Can anyone help

Thanks

Scio
scio2

There is the possibility that this is due to adsorption to the seal material. Waters is offerering different seal materials to solve such problems.

thanks Neue, I will look into this
Scio
scio2

Also make sure your needle wash is actually working. You should get liquid spitting out of the yellow line when you are priming your needle wash.

Also the seal wash frits can become contaminated. Try replacing them. There are two, one on the bottom and one on the top of the injector seal pack.
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry