In order to meet the TNI requirements for 8260, and 624; we have to run 12 sets of PT samples twice a year (you have to pass for each matrix and each compound). This is costly, time consuming and in my opinion not necessary. If you are able to successfully analyze one PT sample of a representative compounds of interest, then that should satisfy the requirement. I like to know what to you think.
For our TNI we run 8260/624 water PT twice per year, and 8260 soil PT twice per year, since there is no soil matrix under 624.
For Drinking water it is a set of Regulated, Unregulated and THM PTs twice per year under 524.2 and a EDB/DBCP twice per year as 524.3SIM.
Not sure if you have other matrix to make it up to 12 sets, but those are ours for the volatiles lab.
The naming is the worst part and it is based on the EPA requirements, not so much TNI. Drinking water branch wants it reported as Methyl Chloride and 1,1-Dichloroethylene and Waste water branch wants it reported as Chloromethane and 1,1-Dichloroethene, as an example.
I am not big on having a lot of laws and regulations but could we at least get one that forces all branches of EPA to use the same naming pattern? They can't even agree on limits for tuning ion ratios in BFB and DFTPP for goodness sake. Then there are the units, our drinking water branch wants everything reported in mg/L so I have a detection limit for Silvex herbicide of 0.0000075 mg/L can't say that isn't difficult to not misplace a 0 on