Column Advice...
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
I am trying to find a column that can get good separation of Argon/Oxygen/Nitrogen and still be able to handle CO2. Any of you guys have any suggestions?
-
- Posts: 12
- Joined: Fri Sep 18, 2009 3:41 pm
Hi GTR, i think this one might work~~
Rt-Msieve 5A PLOT
Here's the link for more info~~~ enjoy
http://www.restek.com/aoi_petro_A009.asp
Rt-Msieve 5A PLOT
Here's the link for more info~~~ enjoy
http://www.restek.com/aoi_petro_A009.asp
-
- Posts: 11
- Joined: Mon Apr 20, 2009 8:05 pm
I've always been told that you can't run CO2 through a Molsieve because it will be retained by the column and poison it.
-
- CE Instruments
We have supplied numerous system to analyse O2, N2 and Argon in CO2. I would be delighted if anyone could find an easy solution. We do this either by trapping and backflush of the CO2 on a Porapak type column , or twin channel with detection of the eluant of the Porapak to another detector both with Molsieve for Oxygen, Argon, Nitrogen separation, or we even set up analysis on a Porapak column in liquid Nitrogen , removing the liquid Nitrogen every few runs and eluting the reatained CO2
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
The Restek solution required a helium ionization detector. A TCD will not work. and if your sample is larger than 5µL you may overload the column.
To separate all these gases you need a MS 13X column at ambient temperatures or a 5A column if you can heat your oven to 200°C.
You can also use a 15 to 30ft Carboxen 1000 column but argon separation would require sub-ambient temperatures if it is possible.
The best ambient or low temperature solution with a TCD is 13X but Argon separation from oxygen is minimal.
Using multidimension chromatography and good packed columns, a 2 meter Porapak Q and a 9 meter MS 5A will separate all the peaks (with a backflush to remove heavies) at 20-30°C.
best wishes,
Rodney George
consultant
To separate all these gases you need a MS 13X column at ambient temperatures or a 5A column if you can heat your oven to 200°C.
You can also use a 15 to 30ft Carboxen 1000 column but argon separation would require sub-ambient temperatures if it is possible.
The best ambient or low temperature solution with a TCD is 13X but Argon separation from oxygen is minimal.
Using multidimension chromatography and good packed columns, a 2 meter Porapak Q and a 9 meter MS 5A will separate all the peaks (with a backflush to remove heavies) at 20-30°C.
best wishes,
Rodney George
consultant
5 posts
Page 1 of 1
Who is online
In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 13 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 13 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science