- Posts: 4
- Joined: Mon Nov 02, 2009 2:17 pm
At my laboratory we currently use a 25mM potassium phosphate buffer, adjusted to a pH of 3 with phosphoric acid. We would like to go away from such a salty buffer as it is destroying most of our stainless steel parts due to leaky pumps, seals, etc.
We use Agilent 1100 HPLC-DADs (there are a few 1200s around but the majority are 1100s). We collect signals at 210, 233 and 250nm so the type of buffer is key to our analysis. As an FYI we are a forensics lab that deals with the analysis of miligram amounts of ilicit drugs.
I've tried ammonium formate and ammonium acetate buffers (both 10mM conentrations and pHs of 3.5 and 4.5 respectivly) I've heard that both of these buffers are great as volatile buffers and are used widely in the world of LCMS. However, both of these buffers have a UV-cutoff that is not acceptable to our standards. We essentially need a buffer that can work around pH 3, be volatile, and have a UV-cutoff of less than 190nm.
After all of that background, here is my question. Would a buffer of ammonium phosphate, with a pH of around 3, be volatile enough so that upon leaks we should not see salt crashing out and corroding our systems?
I've done a lot of searching online and the only hits that I come up with pertaining to ammonium phosphate have to do with protein analysis at pH 7.
Any help would be much appreciated.
Thanks!