You will probably have to contact Agilent technical service directly for the details of the tests you wish to do. Naturally they will want to send out one of their service engineers a major cost but perhaps you will be able to get them to give you the details you require. Good luck

Well yeah, of course they'll send some one out and it costs money. Its a service provided by Agilent and the software, methods etc are proprietary

. You can always do the qualification yourself and deal with the QA/regulatory implications. Keep in mind that there are other benefits that come with getting an official Agilent OQ, for example: the way we run the protocol creates an audit trail for all data, methods, and sequences. All original data/methods/sequences are also stored with the results for future review.
DISCLAIMER: I am an Agilent service provider and the majority of my work is running qualifications, everything I say here is subject to guesstimation and is not The Law. Contact Agilent, get the ACE EQP for your system to get some official information. If you need to make changes to the limits or configuration that can all be discussed with the engineer before you approve the protocol. If I was going to do your system this is what the setup would be:
Channel 1: ALS -- PP -- TCD
Channel 2: HS --- SS -- FID
Channel 3: ALS -- SS -- ECD (depending on how you use the ECD, if you use it with the headspace I'd test it with the headspace attached) *see note at bottom
Tests to be run on the headspace include heated zone accuracy for the oven and sample loop, solenoid valve function test to check for leaks with the vent and pressurization valves. A series of injections would be made to confirm reproducibility of injection area counts and retention times. Another injection is made to check for carryover/contamination.
For the GC:
Inlets are leak tested and the pressure accuracy is also tested. Oven temperature accuracy is checked at a low and high set point, and stability is checked by comparing drift at an isothermal setting. Detector flow accuracy would also be checked for the FID and TCD. I haven't run an OQ on an instrument with an ECD in a little while but I don't think it has a flow test.
Injection precision is checked for the ALS by making another series of inections, signal to noise would be tested for FID/TCD using both the ALS and headspace and then the ECD would also be tested (probably with the ALS but this is up to the customer). A noise and drift test would be performed for all detectors as well.
* Note that a separate charge is included for testing the 3rd detector, linearity is also an additional test which requires an extra charge.
Let me know if you have more questions.