WHat solvents/ Mobile phase affect the performance of RI det

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I just want to know that what solvents may affect the stability of RI detector and what other parameters should be kept in mind while working with RI Detector?

Premix the mobile phase and pump from one channel; best to bypass any mixing valves and go straight to the pump.

As stated, premix your MP. Second, I've found that ACN based MP's tend to stabilize more quickly than MeOH based MP's, at least in RP applications. Why this is, I'm honestly not sure. Other things: Use a PEEK fitting on the RI inlet (lower thermal conductivity than SS) and at a minimum keep the column insulated if you don't have a column oven. If you do have a column oven, set it at the same temperature or slightly warmer than your RI cell temperature. Other than that, be patient. Purging the cell well then recirculating your MP overnight is a convenient way to come in to a fully equilibrated machine in the morning with a nearly full MP reservoir.
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry