Accela 1250 Pump blocked

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi there!
I had a problem with the quaternary pump on my LC system. The system was shut down for almost 4 months and was last closed with MetOH 50%.
I wanted to turn it on but the pumps were blocked. Reading the manual I thought I should remove the LDA and clean the check valve.
Would this be a solution?

Thanks.
won't hurt... Also, make sure your proportioning valve works - that all of the solenoids are opening when they should.
Thanks,
DR