fame analysis
Posted: Mon Mar 28, 2022 7:39 am
by kaliakon
I am trying to analyze fame (gc-fid) with sp2380 (L×I.D. 30 m × 0.25 mm, df 0.20 μm) gc column. When the supelco 37-fame mix is analyzed, the separation between C18:1 trans & cis is perfect. But when I analyze olive oil, I cannot see the trans isomers. I tried higher split ratio, stable temperature at the time that these fame are observed, but with no success.
Any ideas?
Thank you for your time
Re: fame analysis
Posted: Mon Mar 28, 2022 8:18 am
by dblux_
Any ideas?
First of all doe's your olive oil contain trans isomers ?
IMHO 30 m column is too short to get sufficient resolution for FAME analysis.
Re: fame analysis
Posted: Mon Mar 28, 2022 8:49 am
by kaliakon
The olive oil that I analyzed is a reference material, so the compotitiion of fame is known. I selected this column because the chromatograph that the company provided, appeared to have perfect separation for the basic cis & trans isomers. It seems that the separation is not so good with matrix samples.
The column is recently purchased, so I am not able to buy a new one.
Do you think there is a way to achieve separation?
Thank you for your reply
Re: fame analysis
Posted: Mon Mar 28, 2022 1:48 pm
by Consumer Products Guy
I also wonder if your sample is low on trans C18:1
Re: fame analysis
Posted: Thu Mar 31, 2022 2:39 pm
by James_Ball
Have you tried to spike some of the reference sample with the trans isomer and see if it appears? If so, then maybe the reference concentration is misstated.