HPLC System Cleaning
Posted: Wed Oct 28, 2009 1:45 pm
Hi Everybody,
I've been having some massive problems with system peaks in my HPLC (A1100):
C18 column
A = 0.2% TFA v/v aq
B = MeOH
35-95% B over 30 minutes
10 minute equilibration
Column thermostatted at 30oC
UV detection at 270nm
The peaks are late eluting at around 20-30 minutes.
I've carried out extensive investigations looking at the water, TFA and MeOH quality. I've used different columns and I've done 0uL blank injections from empty vials with no septa to rule out vial artefacts. None of this has allowed the cause to be identified.
Moving to different a system gives a different profile of artefact peaks. This suggests that it's system contamination and that there may be similar, but not identical contamination on each system. Moving to a system in a different lab gives clean baseline.
I've flushed the system with neat MeOH, MeCN, IPA and with 1/1/1 Cy/MeOH/IPA. I've also flushed it with water and with 10% HCL and 10% NaOH. None of this made a significant difference.
I've run blank injections with the column switch valve and peltiers isolated and also with the injector isolated (ie: plumbing the pump outlet directly into the column) and the peaks were still seen.
I've swapped my aqueous and organic lines round and I've also cleaned my mobile phase reservoirs to ensure there's no organic contaminants in there.
I can only think of a few further things:
1) Replace solvent inlet glass frits/PTFE solvent lines
2) Contamination within the degasser (possibly microbial?)
3) Flush with an even stronger organic solvent (DCM)
4) Passivate/flush with HNO3 aq
The systems have been used with phosphate aq at pH 7 in the past so microbial growth is a real possibility.
Any thoughts?
I'd really appreciate any advice on this. Also, if anybody knows of a proven protocol for cleaning an A1100 up from either organic or microbial contamination or both I'd be very grateful too.
Thanks in advance!!
C18 column
A = 0.2% TFA v/v aq
B = MeOH
Any thoughts?
Thanks in advance!!