Is spike recovery could be best way to show the accuracy of the analytical method or other better way if yes let me know. The way of showing accuracy could be different by different chromatographic method like SEC,IEC,RP etc.
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Depends on what’s the goal of the analysis. If determination of a compound (drug substance for instance) in a complex matrix (a pharmaceutical formulation for instance) is the aim, then spiking is probably the most convincing method – especially if the separation is less than perfect! Otherwise (if analysis of a single compound (“pureâ€
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- grzesiek
"The way of showing accuracy could be different by different chromatographic method like SEC,IEC,RP" - the same, reading some guidlines can give you some more understanding, start with ICH Q2
"Is spike recovery could be best way to show the accuracy of the analytical method" - recovery study is probably most widely used, what is best for you may not be best for someone else
"Is spike recovery could be best way to show the accuracy of the analytical method" - recovery study is probably most widely used, what is best for you may not be best for someone else
-
- Posts: 17
- Joined: Fri Jun 12, 2009 5:42 pm
My knowledge is with drug formulations (mainly syrups and suspensions). The best way I have found is to spike placebo with known concentrations of the API (usually at 80%,100% and 120%) in triplicate prep and inject in duplicate. I will typically put a 97-103 or a tighter 98-102 recovery spec for the accuracy
"I have not failed. I have just found 10,000 ways that won't work" Thomas Edison
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
I do very similar for the Accuracy requirement. Actually, “METHODS, METHOD VERIFICATION AND VALIDATIONâ€My knowledge is with drug formulations (mainly syrups and suspensions). The best way I have found is to spike placebo with known concentrations of the API (usually at 80%,100% and 120%) in triplicate prep and inject in duplicate. I will typically put a 97-103 or a tighter 98-102 recovery spec for the accuracy
6 posts
Page 1 of 1
Who is online
In total there are 19 users online :: 1 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 18 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 18 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science