Can anyone advise on why we get negative peaks on our solvent front with a reverse phase assay method when the sample solvent is mobile phase. Wavelength is 210nm and mobile phase is 0.4% THF, 1.0% Isopropyl Alcohol, 98.6% water plus TFA to pH 2.6
Is it the TFA that causes this problem ?.

Any hints would be appreciated. Thanks.

Sarex
Sarex-
Log in to www.waters.com. On their home page search window type " HPLC Troubleshooting". The second result that comes up will be for their Information Center for a publication by Dr. Uwe Neue (#WA 20769). You can download the whole thing in pdf. There is a chapter in there taht describes your situation very closely. Hope it helps.
-Gold Leader

btw- I'm not associated with Waters in any way...
2 posts Page 1 of 1

Who is online

In total there are 36 users online :: 3 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google Feedfetcher, John Guajardo and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry