I have to do my methanol and phenol calibration at different split ratio (100:1, 50:1, 20:1, 10:1) . But I found the response factor does not change linearly as the split ratio change. The following are my calibration equations at different split ratio (Y is peak area, and X is concentration)
methanol
split 10 y=1125x
split 20 y=412.19x (For methanol, you can see, as split ratio decrease from 100: 1 to 50:1, the
split 50 y=165.61x response factor increased from 28.275 to 165.61 (around 7 times)
split 100 y=28.275x
phenol
split 10 y=2925.3x
split 20 y=1576.3x
split 50 y=477.87x
split 100 y=105.27x
For the calibration curve at each split ratio, the reproducibility is good, and the R^2 is >0.98.
I don't what's the problem? calibration solvent? or improper column length in the injector port so that too many sample go to the column instead of the split vent?
Thanks!