potassium sorbate with RID
Posted: Tue Oct 27, 2009 5:34 am
by isuphysics
I am trying to test for potassium sorbate. Is it possible to use RID or do I have to get an UV? Any help is greatly appreciated I'm using the following column: Nucleosil 5 C18 120A.....
Posted: Tue Oct 27, 2009 9:33 am
by virtu
UV - if concentration <500ppm
Posted: Tue Oct 27, 2009 2:55 pm
by Consumer Products Guy
I've assayed potassium sorbate by GC after acidification to sorbic acid.
HPLC with low wavelength UV can be used for a lot of stuff - provided that the matrix doesn't interfere at the low wavelength.
Posted: Tue Oct 27, 2009 8:37 pm
by Bryan Evans
Below is potassium sorbate on Unison UK-Amino & Unison UK-Silica:
http://www.imtaktusa.com/site_media/fil ... TI482E.pdf
Posted: Tue Oct 27, 2009 11:16 pm
by isuphysics
Thanks for all for your help..
Posted: Wed Oct 28, 2009 8:32 am
by XL
As suggested above, use UV if only sorbate is the analyte of interest.
If you need to detect and quantify potassium in the same analysis for any reason, you can use Acclaim Trinity P1 from Dionex (http://www.dionex.com/en-us/webdocs/707 ... 239-01.pdf) and a Corona CAD or an ELS detector. Please contact me if you are interested.