Peak Purity...any way around it?..if it fails!!
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
Im quantifying a peak for assay and the current developed method we have., has a small degradant on the tail of the peak causing the peak purity to fail!!...is there any way i can justify this , thus allowing me to continue using this method
-
- Posts: 50
- Joined: Tue Oct 06, 2009 2:09 pm
If your software can integrate the contaminant as a skimmed peak then you might be able to justify it.
-
- Posts: 239
- Joined: Tue Nov 09, 2004 3:06 pm
If the peak has an impurity peak on its tail it IS NOT pure. It is simply not separated. And if it is not pure and not separated there is no sense in using peak purity.
If the impurity was on the far tail and you could convince your integrator to make a valley-to-valley integration the peak might turn out pure.
What do you do in assay? Do you intergrate with the impurity peak?
If the impurity was on the far tail and you could convince your integrator to make a valley-to-valley integration the peak might turn out pure.
What do you do in assay? Do you intergrate with the impurity peak?
-
- Posts: 35
- Joined: Tue Feb 03, 2009 11:25 am
Your question worries me a little. The method was setup to check peak purity for a reason... and the reason was just triggered. As already stated you really have to separate it so the integration does not include it in the main peak. in this scenario it may be that your peak purity settings are a bit too sensitive and you need to reset the threshold (noise reject) value.
Always remember that in UV peak purity you can never be sure of purity (co-eluant may have spectrum very close in appearance to main peak). Impurity detection can be fooled by noise at front and tail of peak. only the right settings allow you to be sure of impurity co-eluant. It's a bit of a subjective art at the best of times.
Always remember that in UV peak purity you can never be sure of purity (co-eluant may have spectrum very close in appearance to main peak). Impurity detection can be fooled by noise at front and tail of peak. only the right settings allow you to be sure of impurity co-eluant. It's a bit of a subjective art at the best of times.
regards, Rob
____________________________________
Rob Sample Consulting - working worldwide to bring analytical expertise to your laboratory.
____________________________________
Rob Sample Consulting - working worldwide to bring analytical expertise to your laboratory.
-
- Posts: 292
- Joined: Wed Jan 19, 2005 2:20 pm
Can you do peak purity by MS instead? Much more valuable in my experience than doing so by DAD.
6 posts
Page 1 of 1
Who is online
In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Google [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science