Agilent 5973 firmware upgrade
Posted: Mon Feb 21, 2022 10:34 pm
by paolo.canu@unipd.it
does anybody know how to fix a firmware that became corrupted because of an unsuccessful update (some *.o files were not copied to the SC)?
The topic appeared in the past, apparently without any final solution.
Symptoms are well known: MS is stuck while rebooting, showing a 'Loading MSD firmware'. I can access the smartcard by telnet/ftp, but apparently, it lacks a "boostHost."
Thanks for any help
PC
Re: Agilent 5973 firmware upgrade
Posted: Tue Feb 22, 2022 9:17 pm
by James_Ball
If you delete the instrument in MS Configuration then reinstall it, does it go through the firmware update process? I believe that is where it usually updates it and maybe that would force it to complete a new update.
Posted: Thu Feb 24, 2022 5:03 pm
Thank you James for your advice.
an MS witout a running FW does not pass the RPC test.
I guess one issue that determined a corrupted FW may be connected with the request of USER/PASS during the update:
C:\msdchem\MSexe\firmware\5973N\msupdate.exe version 3.98
Can't get current MSSC revision. MSSC not active...
Current OS version: 1.2/22T
SC3type = SC3Plus
Update MSSC version: 5.02.07
Update OS : skipped
MSD SmartCard at 10.1.1.102 will be updated
!!>>DO NOT DISRUPT POWER TO MSD DURING THE UPDATE.<<!!
OK to proceed (Y=YES): Y
Clearing /flash/:
Found an existing BOOT.ASC file.
rm "/flash/MSSC.O"
value = 0 = 0x0
-> rm "/flash/LUI.O"
value = 0 = 0x0
-> rm "/flash/STARTUP"
value = 0 = 0x0
-> rm "/flash/httpd.o"
value = 0 = 0x0
-> rm "/flash/wdsmall_run.o"
value = 0 = 0x0
-> rm "/flash/BLINK.O"
value = 0 = 0x0
Sending 10.1.1.102:/flash/MSSC.O
USER and PASS required
Sending 10.1.1.102:/flash/LUI.O
Sending 10.1.1.102:/flash/STARTUP
Sending 10.1.1.102:/flash/httpd.o
Sending 10.1.1.102:/flash/wdsmall_run.o
Sending 10.1.1.102:/flash/BLINK.O
Initiating flash burn.
diskFlash
erasing bank 1: sector 0
programming data...
flash update complete...
value = 0 = 0x0
Code download complete.
Power cycle instrument to complete update process.
Posted: Tue Mar 01, 2022 8:34 am
SOLVED.
INstalled Chestatin E.02.02 and the firmware was regularly updated to 5.02.14.
The USER and PASS required remained, but the procedure completed successfully