Charged Aerosol Detector Contamination
Posted: Tue Feb 01, 2022 8:07 pm
by chromacolor
Hello, I was wondering if anyone else has ever seen an orange/brown oily substance coming from the exhaust tube out of a CAD? It is a thick substance and is only seen on the outlet, the inlet tube is clear with no trace of the contaminant. We checked the purity of the N2 source and it is 99.9%, does anyone have any other theories what this could be?
Re: Charged Aerosol Detector Contamination
Posted: Thu Feb 10, 2022 3:26 pm
by RHWoodman
Hi, Chromacolor,
I have no idea, but we have the same issue with one of our two CA detectors (both Thermo Fisher Corona Veo). Ours isn't brown, it's more yellow-orange, very clear, and thick. We put in a request for help with Thermo but haven't heard anything back aside from an acknowledgement that they were investigating. If you learn anything, I'd love to know what it is.
Best,
Robert
Posted: Thu Feb 10, 2022 4:30 pm
by chromacolor
Hi, Chromacolor,
Best,
Robert
Hi Robert,
What type of analysis do you run on the CAD detectors? DO you happen to run lipids?
Posted: Thu Feb 10, 2022 5:31 pm
by chromacolor
Hi, Chromacolor,
Best,
Robert
Hi Robert,
Also, what types of MP do you typically run with? Do you happen to know what year the CAD was purchased? I am trying to determine a common cause that may help us identify the cause.
Posted: Sat Feb 12, 2022 6:11 pm
by Vlad Orlovsky
Most likely this comes from your air supply line. If you don't have filtering system or traps, this might be coming from your compressor.
Posted: Sat Mar 19, 2022 12:58 pm
by BostonFSE
Are you running lipids with THF or ipa as your mobile phase?
Posted: Wed Apr 27, 2022 9:33 pm
by RHWoodman
Hi, Chromacolor,
Best,
Robert
Hi Robert,
Hi, Chromacolor,
I apologize for the late reply. We use the Corona Veo CAD for running low M.W. PEG derivatives. Our mobile phases are typically acetonitrile and water with 0.1 vol% in each. Our nitrogen is supplied either by a nitrogen generator from Peak Scientific or a nitrogen cylinder (grade 4.8 or higher) from Praxair.
Best,
Robert
Posted: Thu Jun 02, 2022 1:57 pm
by chromacolor
We were running with different MPs including a combo of IPA, ACN, MeOH, and water containing TEA, TFA, and FA in different combos.
Posted: Mon Oct 24, 2022 7:54 pm
by chromacolor
Yes
by chromacolor
Yes