Page 1 of 1

manual hplc??

Posted: Wed Oct 21, 2009 8:37 pm
by che313
it could be to run hplc without chemstation?
we have the isocratic bump and hplc column which are connected to icp-ms

so i'm thinking to use the column in separation then collecting the fraction and analysing them later by gc-ms or tlc

Posted: Thu Oct 22, 2009 3:47 pm
by elsico
you can use agilent handheld controller if you think thats enough
http://www.chem.agilent.com/Library/use ... _ebook.pdf

Posted: Thu Oct 22, 2009 4:30 pm
by Consumer Products Guy
"you can use agilent handheld controller ....for 1100 modules....if you think thats enough
http://www.chem.agilent.com/Library/use ... _ebook.pdf"

And 1050 modules can be run just from their front panels.