- Posts: 22
- Joined: Tue Dec 07, 2004 1:39 pm
I took a 4 or 5 month break from some method development work to do some survey work.
Upon returning, I discover that I had lost response and sensitivity on a fluorescence detection method, the sensitivity was also lost (i.e. the slope of the calibration curve).
I have ruled out bad preparation of the standards and reprepared them fresh from a stock solution. The stock solution was verified using UV-Visible Spectroscopy.
I have ensured that the gain was not inadvertently changed. I have performed the Raman S/N test.
The analyte I am looking at is Ochratoxin A with an excitation wavelength of 333 nm and an emission wavelength of 460 nm.
The detector is a Waters Acquity Fluorescence detector.
I will also mention the flowcell appears to have leaked at some point as there is encrusted material on the exterior, however it does not seem to be leaking now.
Our lab would rather not spend $4200 until we are sure that it is in fact the flowcell.
Would anyone have any suggestions as to how to test this flowcell or why the response and sensitivity were lost.
Cheers,
Tom