Dear All,

We are using Agilent 6890 gas chromatograph and agilent mass detector for a halogenated monoterpene study. After a few runs, the GC-MS gave stick-like chromatograms. The image of the chromatogram is here: [/img] http://studwww.ugent.be/~dpayo/bad_chrom.jpg

On the lower left is a closer view of one peak and on the lower right is the mass spectrum.What could be the problem? The person assisting me could not explain why. If I need to provide some more details to help diagnose the problem, let me know.

Thanks for any help.

DPayo

Dpayo,

I cannot see the image.

Does your MS pass the tune?

Regards,
bhuvfe
Dear All,

We are using Agilent 6890 gas chromatograph and agilent mass detector for a halogenated monoterpene study. After a few runs, the GC-MS gave stick-like chromatograms. The image of the chromatogram is here: [/img] http://studwww.ugent.be/~dpayo/bad_chrom.jpg

On the lower left is a closer view of one peak and on the lower right is the mass spectrum.What could be the problem? The person assisting me could not explain why. If I need to provide some more details to help diagnose the problem, let me know.

Thanks for any help.

DPayo
I can not see the chromatogram neither. Anyway, you can check the ion source. Maybe the sample is not clean and it messed the source. I just guess.

Dpayo,

I cannot see the image.

Does your MS pass the tune?

Regards,
bhuvfe
Thanks for the reply, Bhuvfe and thohry.

I was just informed that the MS filament needs to be changed.

I noticed that the link only works with Firefox.

I could see indeed the image with firefox.

What you see is mainly noise. Do an autotune and see what happens (and maybe switch to the other filament)

You can get a similar "chromatogram" with a very low voltage on the electron
multiplier.

I could see indeed the image with firefox.

What you see is mainly noise. Do an autotune and see what happens (and maybe switch to the other filament)

You can get a similar "chromatogram" with a very low voltage on the electron
multiplier.
Bhuvfe,

Thanks for the info. I'll discuss this with the technician helping me. I'm just starting to learn this instrument.
6 posts Page 1 of 1

Who is online

In total there are 115 users online :: 0 registered, 0 hidden and 115 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 115 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry