I contacted Waters, but their answer is intriguing. Any suggestions would be highly appreciated!
-------------- Waters' answer
it's quite possible that the PDA model type is the issue here. There is a large enough difference between how the 2487 UV detector works and your model (the 2487 has the ability to do some filtering since it uses a reference) but in your case, what your detector sees is what you get. The bandwidth of the cell and the function of the diodes all play a role with what you're seeing and it can't be altered to achieve a similar result as in the 2487. In addition, the UV cut off for ACN is below 200 but close enough to the regions of your interest that the noise generated can't be filtered out on this detector type.
However, there's one thing you might be able to try that could help the situation. If you're using 100% ACN and 100% H2O, you could try to use a premix of say 95% ACN with 5% H2O and 0.1% TFA for solvent A and for solvent B you could use 95% H2O with 5% ACN and 0.1% TFA. This composition seems to work better than the straight 100% composition in decreasing the amount of out-gassing to the cell. If this is something you think you could use in your analysis, I'd give it a try. Otherwise if you need to use ACN, you might have to use a different detector type.
------------------