Page 1 of 1
Quadratic Fit for Anions by Ion Chrom [June 25, 2004]
Posted: Tue Aug 24, 2004 7:35 pm
by admin
By Vangelis on Friday, June 25, 2004 - 06:37 am:
A colleague with whom we're cooperating on IC analysis, informed us that (according to Dionex-stuff) when one does Anions analysis with IC, using an eluent consisting of carbonate, they should use quadratic fit. Would it be possible that anyone explains why could/should this happen (or paste a link either here or on an email to me of a relative document?) I've been searching the website of DIONEX but haven't found any document that suggests this for this specific application.
Thanks in advance,
By Chris Pohl on Saturday, June 26, 2004 - 03:59 pm:
Vanglis,
The most detailed reference is: M. Doury-Berthod, et al. Analytical Chemistry, Volume 57 (1985) page 2257. You'll also find information regarding this aspect on pages 277-279 of Paul Haddad's book: Ion Chromatography, Volume 46 of the Journal of Chromatography Library.
Basically, the weakly acidic character of the carbonic acid background induces a minor amount of nonlinearity. You don't really to worry about a quadratic fit unless you are dealing with a very broad calibration range.
By Vangelis on Monday, June 28, 2004 - 12:31 am:
many many thanks Chris,