by 
sabine » Thu Oct 15, 2009 7:52 am
													
 
					
						Hi!
My guide says a fresh column behaves like that and the protein will bind more as the column grows old.
I have never heard of nor experienced something like this with the ion exchange resin we use in pilot/large scale production. Q Sepharose is one of those resins...
If I understood correctly: you equilibrate the column with the buffer mentioned, then load the column (do you use the same buffer for the column load?), then wash the column with that phosphate buffer (then you loose 70%), finally you elute the remaining 30% with the same buffer (containing NaCl).
I would simply say, the wash buffer is not suitable. Either its conductivity is too high (
very unlikely, 10 mM is really low) or the ph is not correct.
So some questions to think about would be:
- are you absolutely sure that your protein is negatively charged at that pH? Otherwise it will not bind correctly. What is its pI?
- do you loose your protein only during washing or is there some loss in the flowthrough during loading, as well? 
-  what sort of guide do you use? GE HC offers pretty good downloadable guides for their resins and on the different separation techniques.
Ah, danko was a bit faster now!
At least we have the same opinion...  