NMP with Head Space
Posted: Thu Dec 09, 2004 12:09 pm
by Basil
Hi all,
I'm interested on an old subject related with analysis of NMP using European Pharmacopeia Method to comply ICH ....
I've seen this old part of conversation related with this compound. Using DMF, DMAc or DMSO I'm not able to see NMP... Our product is not soluble with water or aquous solvent..... I don't know where I can find ethylhexanol as solvent to try to use it.
Basil
__________________________________________________________
By Anonymous on Friday, July 2, 2004 - 05:17 am:
yes, i know, that diluting works fine.
But, again, I cant find NMP. I tried H20, THF, and H20/PEG-1:1-mix as solvent. Maybe someone has an idea why NMP can't detected when it's diluted. Ethylhexanol works well.
Instrument:
GC6890, HSS7694
Inlet:
EPC-Split-Mode, 250°C, Pressure:12psi, Total flow: 63 ml/min, Split ratio 50:1, Gas saver:20ml/min
Column:
HP5-MS, He-flow: 1 ml/min, v=25cm/sec.
Oven:
60°C inital, 5°C/min till 16min, 8°C/min till 25 min, 15°C/min till end
HSS:
Vial temp: 100°C, loop: 110°C, tr.line:120°C
Vial equilibration time: 5min (2.5µl - 45min at 1ml), Pressurizing time: 0,2min, Loop fill time: 0,1min, loop equilibration=0,02min, injection time=0,1min
NMP by headspace
Posted: Thu Dec 09, 2004 1:45 pm
by chromatographer1
What chemical is abbreviated as NMP ?
Posted: Thu Dec 09, 2004 3:56 pm
by Basil
NMP: N-methylpyrrolidone
NMP
Posted: Thu Dec 09, 2004 6:17 pm
by chromatographer1
This solvent has a low volatility and requires a very inert system to be detected by GC.
If present at low levels it would tend to be adsorbed by the rotor/steel transfer line of a headspace unit.
This is one analyte that should be tested by a PE headspace analyzer or by direct injection methodology. A fixed loop or sample valve system will perform this analysis with difficulty.
Expect temperatures higher than the boiling point of water in order to determine this solvent accurately on a very inert thick film capillary column.
If I can come up with direct parameters I will share them.