Is it safe for column lifetime using phosphate buffer at pH 10? Have you try other type of buffer, i.e. TEA (pKa= 10.7), 1-methyl-piperidine (pKa = 10.3 ), glycine (pKa = 9.We use for assay and dissolution test (drug analysis): TSK OD-2PW column, TosoHaas, 0.05 M K2HPO4 (pH 10) :AcN - 45-55, detection at 206 nm. Solvent for samples - mob. phase.
Thanks, Steve. The problem is the detector: electrochemical, guard cell at +1V, dual electrode cell at +0.7 and +0.8V (as in USP). Could the method be run in most common detector (UV-Vis/PDA)?Here is a procedure for the product in human serum. The same parameters will work for your analysis.
Found at https://www.macherey-nagel.ch/web/MN-WE ... =HPLC00361