Advertisement

Accurate Mass APCI

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hi all

have any of you tried doing Accurate mass measurements using APCI as ionization technique?
if so what reference compound works?
i've tried PEI, sodium TFA (not great I know) and PEG.
PEI seems to give the best signal but lacks reference peaks in the low mass range...

any ideas would be appreciated

kind regards

MassSpecM,

I know Agilent sells a tune mix for APCI (but I have no direct experience with it).

http://www.chem.agilent.com/Library/msd ... nglish.pdf


Best regards,

thanks a lot!

i'll try this and give feedback!

MassSpecM,

I know Agilent sells a tune mix for APCI (but I have no direct experience with it).

http://www.chem.agilent.com/Library/msd ... nglish.pdf


Best regards,
Yes this is the same standard solution you can use with ESI source.
Dilute it 1:10 with ACN and inject it.
The conc. solution is quite stable, you can store it at room temp without any degradation. The diluted one is better preserved at 4°C.
Samuele Pedraglio
Developability Dept.
NiKem Research S.r.l.
Italy
4 posts Page 1 of 1

Who is online

In total there are 78 users online :: 0 registered, 0 hidden and 78 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 78 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry