The instrument won't go to ready. It's reading 0 pressure, although it has 75 psi of head pressure. We're reading 28 ml/min through both TCD channels with a flowmeter, so I don't think we have a serious leak.

We've done the EPC calibration through the front panel three times, no help.

We've tried to do the septum purge calibration, but can't get past the input pressure line, since whatever pressure we enter, it doesn't reach (or it thinks it doesn't reach).

We've rebooted several times (if all else fails, reboot!) and this didn't help.

We haven't leak checked the column.

The septum is replaced ea 90 days, although this is a headspace method, so the septum is never pierced. Seems unlikely that we have a serious septum leak.

We're waiting on a response from Varian. Any other ideas?

if you have headspace try to bypass system and run gc standalone. Maybe the leak is in headspace sampler (broken transfer line, etc)

good luck
------------------
daniele
------------------
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry