If you are trying to get better sensitivity for higher mass ions, changing tune, as mentioned by chhubert is particularly helpful for quadropole instruments. And that answer prompted me to carry that suggestion a bit further:
If you are using a quadropole or a low resolution TOF (integer mass resolution) addjustment of mass acquisition window may help. For low mass ions, thermal broadening of the mass and issues related to mass defect can be ignored. But, as you get to higher masses, the thermal broadening of the masses can cause a portion of the mass for one ion to fall into the acquisition window for a neighboring mass. As you get into this range, mass defect becomes a factor for sensitivity as well. Because masses of nuclei are not integer values - and differ based on which atom the exact mass of a molecule of one formula will differ from the exact mass of a molecule of a different formula -- which we use in high resolution mass spectrometry to obtain the molecular formula for an ion. In low resolution mass spectrometry, this moves the apex of the mass peak within the acquistion window - and can force the tail of the distribution out one side or the other of the acquisition window - with resulting loss in sensitivity. When the mass is large enough in a low resolution mass spectrometer, the tails fo the energy distribution of the ion fall out both sides of the acquisition window and may even be counted in the abundance of the neighboring masses.
To inscrease sensitivity for unit mass resolution instruments for high mass ions, you can adjust the mass acquisition window to account for the mass defect - and depending on possibility of interference from signals at the neighboring masses, you may be able to widen the window to catch both sides of the energy distribution of the mass peak.
All of this depends on the particular instrument you are using - and the control you have over acquisition parameters.