There are actually two points in this solution.
First thing I observed that carrier gas is connected to O2 / H2O filter which was way to saturated (to old, spent out), but I found a new one and replaced it. Right after that water level was "just" High (not massive anymore).
Second thing I did was, of course, "Baking out"! I used the temperatures of Trap, Manifold & Xferline, 170, 80, 220 °C respectively and left in "Baking mode" for I think 14-15 hours over night, after that all selftests went fine and I was able to do first measurement after quite some time, which was great.
Thanks bhuvfe for help.
Before this system I had hp5980, i think, wich broke down, so I need to learn again how to work on varian.
Next problem now is to read result after the analysis.... I couldn't get it to compare analysis results with nist database, I'm not sure what to do first after analysis is done to compare it with database........ I'll all of it again in a two week time.