Hi
In other words,I wonder if there is acceptance criteria about validation of analytical procedure,forexample,the data of HPLC method system suitability test,accuracy,precision etc .
From a pharma buisness regulatory viewpoint there is not much in terms of acceptance criterias in the ICH guidelines. Some individual agencies may have a few expectations but to my knowlegde only APVMA (Australia) has set accetance criterias in their guideline for accuracy and precision depending on nominal % of target analyte, see page 5 in the link below.
http://www.apvma.gov.au/guidelines/down ... ethods.pdf
As Tom mentioned, the complexity depends on the purpose and field of work, guidance documents are good but may not "fit all" or reflect the current general "standard" of our instruments. For instance, as for precision in the pharma buissness a RSD of max 2,0% is the general rule, but I would say there is something wrong with "mine" instruments if I obtained a RSD of 1,0 to <2,0% in a general assay or impurity method, as such we tend to set the system suitability critieria to not more than 1,0%.