Need help with separating nitrate and nitrite usin CBM20A?
Posted: Thu Oct 07, 2021 6:49 pm
by skiran2888
Dear all,
I was trying for the past several months to separate nitrate and nitrite from the water sample. I am using the CBM 20A HPLC system enabled with the c-18 column. Kindly, advise me with the best possible methods.
Thank you for your time and effort.
Re: Need help with separating nitrate and nitrite usin CBM20A?
Posted: Fri Oct 08, 2021 2:20 pm
by H.Thomas
Hi skiran, are you serious about that question? You need an anion exchange column for anions - no chance with a C18.
Posted: Fri Oct 08, 2021 2:56 pm
by vmu
Ion-pair chromatography...
Posted: Mon Oct 11, 2021 1:41 pm
by Vlad Orlovsky
You can try something with RP, but I am not sure how well it will work. Most of the RP columns have some residual silanols. You need to inonize these silanols by going to higher pH, something like 6-8. You will be exploring ion-exclusion mechanism. Your flow rate needs to be 25% of you usual rate, but you are expanding this region 4 times because of the lower flow rate (6.5 min for 150 mm column, 10 min for 250 mm column). Nitrate/nitrite will elute BEFORE void, but might be separated. I don’t know how much organic will give you better results, try to explore 20-70%. Use 10-20 mM buffer.
Alternatively you can use one of our methods:
https://helixchrom.com/compounds/nitrate-ion/
Contact me through our website “Contact Us” if you have any question or would like to discuss other options.
Posted: Mon Oct 11, 2021 5:52 pm
by Vlad Orlovsky
You can also try to use a bare silica column with the same approach, but if you start going up in ACN (to 70-90%) you will start seeing effect of HILIC (watch for solubility of salt buffer in high ACN mobile phase) and your compounds might start retaining and separating by combination of HILIC and anion-exclusion mechanisms. Not sure about peak shape, probably is not going to be great.
Posted: Tue Oct 12, 2021 9:56 am
by dblux_
Dear all,
How are you going to detect mentioned anions ?
Posted: Fri Nov 12, 2021 1:14 pm
by skiran2888
I have followed this article reported the Simultaneous determination of...
https://www.ajol.info/index.php/sajc/ar ... le%20phase) Pl. let me know, the same can be followed for the qualitative measurement of nitrate and nitrite ...
thanks
Posted: Fri Nov 12, 2021 1:36 pm
by skiran2888
Dear all,
How are you going to detect mentioned anions?
We are using SPD-20A UV-Vis detector attached to CBM 20A
Posted: Mon Nov 15, 2021 11:01 am
by H.Thomas
thanks
Dear Skiran,
in this article they did not use a C18 column - Synergi Polar is reversed phase material, not C18, but an ether-linked phenyl phase with polar endcapping. Apparently the polar endcapping has some anion-exchange capability. If you look at the chromatograms in the mentioned article, you will find that this is at best a quick-and-dirty method. NO3 has a k<1.
If you really want to use a C18 column, you will need an ion-pairing reagent. But I would suggest to think about buying a diffent column.