I'm interested on an old subject related with analysis of NMP using European Pharmacopeia Method to comply ICH ....
I've seen this old part of conversation related with this compound. Using DMF, DMAc or DMSO I'm not able to see NMP... Our product is not soluble with water or aquous solvent..... I don't know where I can find ethylhexanol as solvent to try to use it.
Thanks
Basil
__________________________________________________________
By Anonymous on Friday, July 2, 2004 - 05:17 am:
yes, i know, that diluting works fine.
But, again, I cant find NMP. I tried H20, THF, and H20/PEG-1:1-mix as solvent. Maybe someone has an idea why NMP can't detected when it's diluted. Ethylhexanol works well.
Instrument:
GC6890, HSS7694
Inlet:
EPC-Split-Mode, 250°C, Pressure:12psi, Total flow: 63 ml/min, Split ratio 50:1, Gas saver:20ml/min
Column:
HP5-MS, He-flow: 1 ml/min, v=25cm/sec.
Oven:
60°C inital, 5°C/min till 16min, 8°C/min till 25 min, 15°C/min till end
HSS:
Vial temp: 100°C, loop: 110°C, tr.line:120°C
Vial equilibration time: 5min (2.5µl - 45min at 1ml), Pressurizing time: 0,2min, Loop fill time: 0,1min, loop equilibration=0,02min, injection time=0,1min