Hi LCHelp,
PLEASE READ THE ENTIRE POST BEFORE MAKING A DECISION, I DISCOVERED A LOT REGARDING THESE COLUMNS AS I DID MORE RESEARCH which is displayed later on in the post!
According to the booklet, the Luna Amino (NH2) column actually has a stability from pH 1.5-11.0! So I was completely incorrect about the pH stability, and rather I was thinking about their cyano columns that are non-stable at higher than neutral pH. Here is an excerpt from their booklet that I found quite interesting: "Column life for most amino columns can be problematic as the amino bonding easy strips off the silica". Immediately after they imply their column is more rugged than other amino columns-- I can't say anything on this front as I don't use this phase regularly. However, I do believe that this disproves the pH theory, but it does not rule out the fact that much of your mobile phase was running at very low pressure with a potentially harmful additive. From many of the vendors I talk to, they recommend to never leave your column in a buffered mobile phase (include bases and acids) and that the pressure is what keeps many of them from degrading in this environment. I'm not sure of the exact reasoning--hopefully someone will pop in and answer this for the both of us, but this could be enough to eventually destroy a column. What I can do is recommend a cleaning procedure directly from the booklet for this column!
Below is their recommended cleaning procedure for Silica-based columns:
-Flow Rates should be 1/5-1/2 of typical flow-rate
-To estimate column volume: V=(pi)r^2*L
V is column volume in mL, r is radius in cm and L is length in cm.
"Reverse Phase Columns (C18,C12...NH2)
Rinse with 10 column volumes each of:
95% Water/5% Acetonitrile
(for buffer removal) (potentially precipitated sugars/salts in your case?)
THF (100%)
95% Acetonitrile/5% Water
Mobile Phase"
^ This is the general procedure for reverse-phase cleaning
NOTE: Reverse the column's flow direction prior to the cleaning!!!! If there is a build-up of some particulate on the end of your column (rather than phase collapse) this could dislodge that blockage and send it down-stream to the waste (disconnect your detector/anything down-stream from column during this exercise).
Another interesting note regarding amino columns that I read is:
"Stationary Phase considerations:
Maintain pH between 2.0-8.0
Use Guard Columns
AVOID ALDEHYDES AND KETONES WITH AMINO COLUMNS"
Since you are working with sugars you're probably dealing with aldoses and ketoses, but this column should be designed for the analysis of sugars so I'm not sure why this comment pops up so late in the booklet. I would probably contact someone from Phenomenex to elaborate on this portion.
Here is what they say regarding column storage (sort of an alternative to having it low-flow over-night every night would be store it in an inert mobile phase and then wash and condition the next morning in your desired mobile phase):
-Column storage conditions affect column lifetime
-NEVER store columns containing buffers or ion-pairing reagents (even though you're not using a buffer, the ammonia hydroxide or whatever base your were using is included in this statement due to the potential of precipitation on the column).
-Flush with at least five column volumes of mobile phase without buffer to remove buffers or salts.
OH GOODNESS I found ANOTHER exception for the Luna Amino column (this thing must be really special: "Exception: (this is under the column cleaning procedure and applies directly to the use reverse-phase amino column usage)
Recommended for cleaning Luna Amino when used in reverse-phase mode:
1. Wash with at LEAST 30 columns volumes of sodium hydroxide pH 11.0
2. Flush with at least 30 column volumes of water (HPLC grade)
3. Re-equillibrate to mobile phase conditions."
Upon further investigation it seems that amino problems kind of suck all-around when it comes to longevity. Here's anecdotal evidence on Research Gate regarding the life-span of these columns:
https://www.researchgate.net/post/Rapid ... PLC-column
Anyway, I hope that second cleaning method is useful and can restore your column. I hope it is just some minor blockage and not phase degradation/collapse! Good luck!