Is there anyone in the list who has experience with carbon based stationary phases. Anyways, certain suggestions would be welcome.
My stationary phase is a commercially available Zirchrom-Carb, which is pyrolytic carbon coated via vapor deposition on ZrO2. Apparently free ZrO2 is not exposed because phosphate elutes very near the injection peak. I am trying to use 1.5 mM tetrabutylammonium hydroxide-1 mM phthalate system at pH -5.5 (the second pKa of phthalic acid) adjusted via titration rather than external acid base. Note I am trying to separation inorganic anions via indirect spectrophotometry. However three ions sulfate ion are causing the most problems:
1. Chloride & HPO4(2-) elute near or in the water dip
2. When SO4 (2-) is present, it disturbs the baseline very near the injection where other simple anions have their peaks.
a. With purely aqueous system, baseline does not remain stable after injection of aqueous anions.
b. With 5% ACN in the eluent, sulfate elutes with the negative system peak.
c. With 10% ACN in the eluent, sulfate causes such a huge disturbance near the negative injection peak of water that sulfate can not be seen, even if elutes. Interestinly system peaks vanish with this eluent but the retention of others anions is decreased very badly that chloride now comes out with the water dip.
Suggestions to increase retention would be welcome and taming sulfate behaviour would be welcome. Note changing pH independently via NaOH or inorganic acids is being avoided because one of the articles said that number of system peaks increase.
Thanks.
M. Farooq