Std curve has gradually gone from linear with r2=0.999 to a curve with high concentrations tailing off over last 6 months. Now a quadratic or log-log fit has to be used. Non-linearity also suddenly became a lot worse after a recent manufacturer engineer visit and MS service.
We are sure this is not due to std preparation, UPLC hardware or columns (we routinely change these). Peak shapes and RT are exactly as they have been when the assay was linear. I suspect the MS. Unfortunately, our analyte is non-UV absorbant, so we can't check this with a UV or PDA detector inline between the UPLC and MS.
I am guessing this is a MS problem, perhaps from dirty ion optics? APCI, ion transfer tube, and first couple of lenses are cleaned/exhanged regularly but hasn't fixed the problem. Has anyone else seen this problem? Or suggest procedures to diagnose the cuase of the non-linearity?
thanks