- Posts: 7
- Joined: Mon Dec 20, 2004 4:04 pm
-The suppressor (CSRS Ultra, 2mm) is leaking internally. This shows by fluids exiting the suppressor at both the eluent and the regenerant ports when injected into the other eluent port with a syringe. I believe this happened because it probably dried up, since it hasn't been used in the last few months. Is there any chance that this is not an irreversible damage? At the moment it is soaked with MSA eluent and water, respectively. Eluent does not flow, since turning on SRS power naturally gives an alarm.
-For hydration and startup purposes I first let the whole system flush with water through the eluent pathway. After a while the backpressure got excessively high, ~1200 psi from the guard column, and >3100 psi from the main column. After reverting to MSA eluent, backpressures normalized rapidly (I first only inserted the guard column, and the backpressures normalized before the eluent could have reached the column). Is there an obvious explanation for this, or should I just be happy that pressures are normal again?
I found this quote on this:
Does the pure water affect some part before the columns?Regarding the suggestion from "Dionex" that running pure water to the resin causes the column to deteriorate (or swell), this is utter nonsense.