TQ LCMS Peak (tr) wander 4 min to 3.5 min
Posted: Thu Sep 23, 2021 8:01 am
Dear all,
I recently joined a bioanalytical team, where LC-MS/MS is used.
I noticed that they run 6 standard curves for each analytical run (using 1 as std curve and 5 other as QC), and normally reject the first 2 (approximately 20 samples), as these are used to calibrate the system, whereafter the system becomes stable. I now also see this phenomenon on the instrument I work on (Sciex 6500+ TQ LC-MS). This affects the reproducebility of my data and reliability on the system.
It is very accepted and normalized in the team that this is how you do things, not only on this instrument. I used to work in another team on Waters Xevo TQ UPLC-MS, and this was not something that I have had problems with.
I then tried to calibrate the system for 30 min prior to running the first samples, but same phenomenon is observed.
Now i also noticed that it takes time for the analyte retention time to settle. It wanders from 4 min to 3.5 min over the course of 40 injections.
I run with MRM, column Acquity UPLC CSH Flupro-phenyl 1,7 µm 2.1x 100mm.
60 degrees celsius for column temp.
I run gradient elution (Mobile phase A: 0.1% formic acid, B: 95% Acetonitrile +0.1% formic acid).
0 min--> 1 min: Fraction B: 10%
1 min --> 4.5 min: Fraction B: 50%
4.51 min --> 5.5 min: Fraction B: 100%
5.51 min --> 7.00 min: fraction B: 10%
Flow: 0.6 mL/min.
Injection volume: 10 µL
Flow: 0.6 mL/min.