Error C0 Corona CAD Detector

Posted: Fri Sep 17, 2021 8:35 pm
by Vlad Orlovsky
Hello,

I have an older version of Corona CAD (ESA/Thermo) which gives me C0 (C zero) error when self testing, which I am unable to find in the manual. Anybody knows what this is?

Thank you.

Re: Error C0 Corona CAD Detector

Posted: Sat Sep 18, 2021 3:16 am
by S_G
https://www.researchgate.net/profile/Is ... ual_EN.pdf

Here's a link from researchgate.net to the manual

Re: Error C0 Corona CAD Detector

Posted: Thu Sep 23, 2021 4:39 pm
by Vlad Orlovsky
I did not see this specific error in the manual, unless it is in the section C of troubleshooting. It states for error 0 - Thermal Organizer not detected

Re: Error C0 Corona CAD Detector

Posted: Mon Oct 04, 2021 6:49 pm
by LCwatcherGE
"C" is a hexadecimal error code that stands for "12". Adding "0" results in Error "120" which is Error 100 (IonTrap OFF --> 24V Power Supply issue) and Error 20 (Pressure Transducer 2 --> Interface Board issue).
A this device is rather old, you'll face the problem to get replacement parts and someone from THermo to service it.
Unfortunately it's better to look for the Corona Veo model.

Greetings

LCWGE

Re: Error C0 Corona CAD Detector

Posted: Wed Oct 06, 2021 1:46 pm
by Vlad Orlovsky
"C" is a hexadecimal error code that stands for "12". Adding "0" results in Error "120" which is Error 100 (IonTrap OFF --> 24V Power Supply issue) and Error 20 (Pressure Transducer 2 --> Interface Board issue).
A this device is rather old, you'll face the problem to get replacement parts and someone from THermo to service it.
Unfortunately it's better to look for the Corona Veo model.

Greetings

LCWGE
Thank you very much for your input.