This is more of a mass spec question than a chromatography question. I'm trying to understand the MS/MS of my analyte, peruvoside. (It's a cardiac glycoside in yellow oleander; MW=548.7.) The analysis is done using ESI, positive mode, mobile phase is 0.1% FA / ACN (95%/5%). I get a nice peak on the chromatograph, which is made up mostly of the M+H ion, but also of an ion at 571, i.e. M+23. Hey, must be a sodium adduct, right? I'm not so sure.
The 571 peak always appears at a level of about 1/3 that of the M+H peak. I have done extractions with and without added NaCl (to enhance phase separation) and there is NO difference in the signal at 571. I would expect the 571 signal to increase, if it's due to a sodium adduct. Where the heck is the sodium coming from, anyway? This "adduct" shows up in my aqueous standards and well as in extracts from plant material.
The MS/MS of the M+H ion (549) consists mainly of ions at 371 (the aglycone fragment), 353, and 389. The MS/MS of the 571 ion consists only of an ion of 411. I would expect the MS/MS of a suspected sodium adduct to resemble the MS/MS of the molecular ion.
Summing up:
1: Wouldn't the amount of a sodium adduct in a MS vary in some way with the presence of added NaCl?
2: Wouldn't the MS/MS of a sodium adduct resemble the MS/MS of a M+H ion, or at least have some ions in common?