lmh » Tue Oct 20, 2009 9:52 am
To go back to the original question, yes, it is possible for two different instruments to give you a different absorbance ratio measured at two wavelengths, because of optical differences in instrument. Two obvious examples:
(1) Bandwidth: obviously if bandwidth differs, so does absorbance. To take the extreme case, if you imagine an absorbance peak that stretches from 300-400nm, if you measure it at 350 and 360nm with a 120nm bandwidth, you'll get no difference whatsoever, but if you measure with a 1nm slit width, you probably will.
(2) Wavelength accuracy: if one instrument is very slightly off wavelength, then the two instruments will agree when looking at the wavelength of maximum absorbance (because here the curve is nearly flat), but disagree when measuring half way down the slope of the absorbance curve (because here there is a big slope).