Page 1 of 1

Separation of D & L phenylephrin

Posted: Sat Sep 19, 2009 5:58 am
by Chandresh K. Soni
how i can Analyze D & L phenylephrin in Reverse phase Chromatography without using Chiral Column on UV Detector. I Have the C-8 , C-18 , Cyano Column For The Analysis.

Posted: Sat Sep 19, 2009 1:42 pm
by Uwe Neue
To separate D/L isomers, you either need chiral additives to the mobile phase or a chiral stationary phase.

The enterprise to do it with additives in the mobile phase is not trivial, and you will be better off taking a pre-cooked method from the literature or from a manufacturer of chiral stationary phases.