The paper presented is not really a Gas Chromatographic method. It is really a Combustion Elemental Analyser but instead of using UVF detection they have used a GC Flame Photometric detector. There is as far as I know no advantage to this method compared to the UVF method as it will be less sensitive and will not result in a lower cost instrument. The selectivity of the FPD may not even be as good as UVF as Nitrogen (used as a Cetane enhancer) is a known interference for UVF
See
UVF elimination of Nitrogen interference
for details of the problem and a solution.
You could use the GC method but Splitless for looking at Sulphur containing compounds by using an FPD after speciation by the GC method. The limiting factor with this is being able to see the very low level Sulphur compounds by FPD once separated. This is why Chemiluminescence is the recommended detection method being much more sensitive than FPD.
Do you have an FPD on your 5890 ? If not then abandon any thoughts now as you would be better off getting a suitable instrument rather than investing in an FPD for your 5890 and hoping that you can meet your detection criteria.
If you want to send me your e-mail address to Richard @ ceinstruments.co.uk I can forward you a few of the pdfs with methods for carrying out this work. (Remove gaps in address