MS on SRI 8610C GC?

Posted: Fri Aug 27, 2021 5:22 pm
by passingthru
Hello everyone, new here.

Has anyone added MS to an SRI 8610C GC?

Re: MS on SRI 8610C GC?

Posted: Sun Aug 29, 2021 5:11 am
by antonk
There are three things the MSD needs from GC
1. The column routed through the hole on GC side. Carrier flow less than 1.5 ml/min
2. Transfer line heater and thermal sensor connected to the spare GC thermal zone.
3. Start line wire (open collector GC input) which should be grounded in moment the GC starts analysis.