1/16" -> 1/32" Tubing

Posted: Thu Sep 17, 2009 11:55 am
by sscala
Is it safe to replace 1/16" SS tubing with 1/32" tubing in connection with conventional HPLC-columns or will the small-diameter-tubing damage the port?

Sebastian

Posted: Thu Sep 17, 2009 1:44 pm
by chromatographer1
If you have the proper ferrules to make a seal I can't imagine how you might damage the column. But why do it? Tubing isn't that expensive. Can you remove all the dead space?

Best wishes,

Rodney George