- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
While working on Polymyxine B Sulphate, according to the Eur Ph method, we discovered that the retention time of the peaks is really unstable. During the sampleset, we can see the retention time is sometimes increasing for a while, and then suddenly decreasing.
First we were thinking that it would be an temperature problem, but after putting the column in a oven set at 30°C, the same issue appeared.
Next thing to know: there are other compounds also injected, and the variable retention time only occurs with Polymyxine, not with the corticoids.
We already tried to switch the method to UPLC (to shorten the run, but also to overcome the problem), but the same problem occurs here.
Method:
4.46g Sodium Sulphate anhydrous / l, brought to pH 2.3 with phosphoric acid. This is mixed with ACN in a composition of 20% ACN and 80% sodium sulphate solution.
Flow 1.0 ml/min (0.4ml/min for UPLC)
Detection 215 nm
Column: Supelcosil ABZ+ 250 x 4.6 5µm (BEH Shield RP18 100 2.1 1.7µm for UPLC)
Column temperature 30°C.
Already tried to put a buffer in the mobile phase, instead of the sodium sulphate, but then the polymyxine looses retention, but also has a variable retention time.
Any ideas?
Ace